Coming back to: Resolution and tailing factor: real acceptance criteria

DO
d.oyelaranTL3Pharmacist1 Jul 2025#1

Posting this under the heading it deserves: Resolution and tailing factor: real acceptance criteria Everything below is what sits behind that.

Reading back through what has been written here about Resolution and tailing factor, three questions come up every time and only one has ever been answered properly.

Listing all three, with what I think the state of the answer is for each, so the thread can start further along than the last one did.

20 likes 13mo
T
ThibodeauTL3Regular4 Jul 2025#2

On integration: where the baseline is drawn matters more than most people realise. On a clean chromatogram with well-resolved peaks the choice is inconsequential. On a chromatogram with a trailing shoulder or a rising baseline it matters. Differences of one to two percentage points between defensible integrations are ordinary.

The general answer and the answer for your case may diverge here.

23 likes 13mo
MR
m.restrepoTL26 Jul 2025 · edited#3
Thibodeau, post #2: On integration: where the baseline is drawn matters more than most people realise. On a clean chromatogram with well-resolved peaks the choice is inconsequential. On a chromatogram with a trailing shoulder or a rising baseline it matters. Differences of one to two percentage points between defensible integrations are ordinary. The… Go to post

Useful. I had the fact and not the reason, which turns out to be the important half.

0 likes in reply to #2 13mo
LO
l.oseiTL27 Jul 2025#4

The practical version of Resolution and tailing factor is three sentences long. The rigorous version is three pages and reaches the same conclusion with the conditions attached.

3 likes 13mo
AA
a.asanteTL29 Jul 2025#5

Narrowing post #4, because the general version has more than one answer.

Having read the whole Resolution and tailing factor thread before replying: the question in the first post has not actually been answered yet, and three of us have answered a nearby one instead.

6 likes 13mo
DB
d.bakkerTL211 Jul 2025#6

Everything in post #2 holds. The case it does not cover is the one I have.

Where the baseline is drawn is the least documented and most consequential choice in the whole determination. Valley-to-valley, tangent skim and forced-to-zero can span a couple of percentage points on the same chromatogram.

16 likes 13mo
PM
p.marchettiTL212 Jul 2025#7
a.asante, post #5: Narrowing post #4, because the general version has more than one answer. Having read the whole Resolution and tailing factor thread before replying: the question in the first post has not actually been answered yet, and three of us have answered a nearby one instead. Go to post

A shallow gradient resolves close-eluting species that a steep one merges. Two honest laboratories running different gradients can report genuinely different purities on the same vial.

Posted with less confidence than the sentence structure implies.

0 likes in reply to #5 13mo
LS
l.solbergTL214 Jul 2025#8

Typical suitability criteria are a replicate area relative standard deviation below about two per cent, a tailing factor inside a defined window, a resolution minimum against a specified peak, and a plate-count floor.

The short version is the first sentence; the rest is why.

1 like 12mo
BA
b.adeyemiTL215 Jul 2025#9
d.oyelaran, post #1: Posting this under the heading it deserves: Resolution and tailing factor: real acceptance criteria Everything below is what sits behind that. Reading back through what has been written here about Resolution and tailing factor, three questions come up every time and only one has ever been answered properly. Listing all three, with what… Go to post

Resolution and tailing factor sits at the boundary between what this community can usefully discuss and what it cannot, and I think it falls on the discussable side, narrowly.

3 likes in reply to #1 12mo
OT
osmolal_tableTL1Member16 Jul 2025#10

Grateful for the specificity. Vague answers to this question are what sent me looking.

11 likes 12mo
NR
n.rahimiTL218 Jul 2025#11

I read the earlier replies on Resolution and tailing factor twice before writing this, because I had assumed the opposite and wanted to be sure I was disagreeing with what was said rather than what I expected.

1 like 12mo
PA
p.amankwahTL219 Jul 2025 · edited#12
p.marchetti, post #7: A shallow gradient resolves close-eluting species that a steep one merges. Two honest laboratories running different gradients can report genuinely different purities on the same vial. Posted with less confidence than the sentence structure implies. Go to post

The arithmetic in post #9 is right; the assumption feeding it is the part to check.

Mobile-phase preparation is a genuine source of between-laboratory variation. Acid concentration and organic modifier both shift retention, and neither is usually specified to the precision that would matter.

I would want to see it done twice before believing it once.

0 likes in reply to #7 12mo
EF
e.ferrariTL220 Jul 2025#13

A note on how Resolution and tailing factor gets discussed rather than on Resolution and tailing factor itself: the confident posts get the replies and the careful ones get ignored, and the careful ones have been right more often.

22 likes 12mo
SS
s.silvaTL221 Jul 2025#14

Following, with nothing to contribute beyond having asked the same thing elsewhere.

10 likes 12mo
K
KLindqvistTL4 Moderator23 Jul 2025#15

Particle size and column dimensions determine what resolution is achievable at all. A 5 micrometre 250 millimetre column and a sub-2 micrometre 100 millimetre column are different instruments in practice.

This is the sort of thing that ought to be settled and apparently is not.

0 likes 12mo
AI
a.ibarraTL224 Jul 2025#16
NL
n.lehtinenTL225 Jul 2025#17

I had written a reply contradicting post #13 and deleted it. Here is what survived.

Where the Resolution and tailing factor discussion usually stalls is that nobody wants to say "I do not know" and everyone is willing to say "it varies". Those are the same sentence with different clothes on.

16 likes 12mo
LD
l.dziedzicTL226 Jul 2025#18

System suitability is not paperwork. If the replicate injections failed, the run did not happen — the numbers from it are uninterpretable rather than approximate.

6 likes 12mo
IR
i.rasmussenTL227 Jul 2025#19
l.dziedzic, post #18: System suitability is not paperwork. If the replicate injections failed, the run did not happen — the numbers from it are uninterpretable rather than approximate. Go to post

Right, and stated more narrowly than I would have dared to state it.

6 likes in reply to #18 12mo
NT
n.torrenceTL3Regular28 Jul 2025#20

A chromatogram image at a resolution where you can see peak shape but not baseline detail is worth having and is not the same as the data. Ask for the integration table if the number matters.

The conclusion is tentative; the arithmetic underneath it is not.

1 like 12mo
M
microgramsTL2Regular30 Jul 2025#21
n.rahimi, post #11: I read the earlier replies on Resolution and tailing factor twice before writing this, because I had assumed the opposite and wanted to be sure I was disagreeing with what was said rather than what I expected. Go to post

Resolution and tailing factor is a good example of a question where the honest answer is boring and the interesting answers are unsupported. I would go with boring.

0 likes in reply to #11 12mo
AA
a.adeyemiTL231 Jul 2025#22

If someone has run Resolution and tailing factor properly I would rather read that than my own reconstruction of it. Posting mine only because the thread has gone quiet.

0 likes 12mo
HO
h.oyelowoTL2Regular1 Aug 2025#23

This follows post #22 rather than contradicting it.

Area percent is a proportion of absorbance, not a proportion of mass. Two species with different extinction coefficients at the detection wavelength contribute unequally to the total, and nothing on the certificate corrects for that.

8 likes 12mo
MR
m.radichTL22 Aug 2025#24

Quietly grateful for the plain phrasing. Not every thread gets that.

20 likes 12mo
SC
sourced_claimsTL3Regular3 Aug 2025#25

Method validation is the demonstration that a method can separate the compound from its degradation products and impurities reliably. A method that cannot resolve an impurity from the parent peak will not detect that impurity.

Adding the caveat now so it does not have to be extracted later.

28 likes 12mo
RZ
ro.zielinskiTL24 Aug 2025#26

The version of Resolution and tailing factor that circulates here is a simplification of a simplification. It is not wrong, but it has lost the conditions under which it holds, and those conditions are where the interesting cases live.

0 likes 12mo
DV
dr.villanuevaTL3Physician5 Aug 2025#27

Column chemistry and particle size: smaller particles (1.7 μm) give better resolution and higher efficiency than larger particles (3.5 μm or 5 μm), at the cost of higher back pressure. Newer methods increasingly use smaller particles.

5 likes 12mo
SG
s.grimaldiTL26 Aug 2025#28
d.bakker, post #6: Everything in post #2 holds. The case it does not cover is the one I have. Where the baseline is drawn is the least documented and most consequential choice in the whole determination. Valley-to-valley, tangent skim and forced-to-zero can span a couple of percentage points on the same chromatogram. Go to post

Post #25 and I disagree about the size of the effect, not about the direction.

Practical answer on Resolution and tailing factor, since the theoretical one is upthread: do the simplest check first, write down the result, and only then decide whether the complicated explanation is needed. It usually is not.

14 likes in reply to #6 12mo
MH
ms_hollowayTL4Mass spectrometrist7 Aug 2025#29

Post #26 is the version of this I will quote in future. One addition.

Genuine question rather than a rhetorical one: has anyone here actually observed Resolution and tailing factor, as opposed to read about it? The thread is long and I cannot tell.

21 likes 12mo
EI
e.iyerTL28 Aug 2025#30
a.ibarra, post #16: Worth stating the null on Resolution and tailing factor before we explain it: the observation may be nothing. That possibility deserves a sentence and usually does not get one. Go to post

Where I part company with post #28, and it is a narrow parting.

System suitability testing: injections run before and during the sample run to establish whether the instrument, column and method were performing when the sample was analysed. If suitability did not pass, the sample results from that run are uninterpretable.

This is the version I would want a new member to read first.

0 likes in reply to #16 12mo