j.cabreraTL212 Dec 2024#61

Where a pharmacopoeial monograph exists, a method that follows it inherits a great deal of assurance. Almost nothing discussed here has one.

0 likes 20mo
F
FFaulknerTL3Regular12 Dec 2024#62

This follows post #61 rather than contradicting it.

Sample stability during analysis is part of validation and is routinely ignored. A preparation that degrades in the autosampler over eight hours produces a sequence-dependent result.

It reads as pedantry until the day it does not.

24 likes 20mo
HN
h.nwosuTL212 Dec 2024#63
t.steenkamp, post #52: I had read the opposite somewhere and cannot now find where, which tells me something. Go to post

Coming back to post #61, because the follow-up matters more than the original answer.

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

Posting it because the silence on this was starting to look like agreement.

11 likes in reply to #52 20mo
TS
taper_shiftTL3Regular12 Dec 2024#64

System suitability is the ongoing evidence that a validated method is still performing. Validation is done once; suitability is done every run, and it is the one that appears on a certificate.

Second-hand, so weight it accordingly.

3 likes 20mo
BA
b.adeyemiTL213 Dec 2024#65

Worth separating two things that post #61 runs together.

Robustness testing deliberately varies the parameters most likely to drift — organic percentage, pH, temperature, flow — and shows the result does not. It is the part of validation that predicts whether a method will transfer.

It is a small point and it changes the answer, which is an awkward combination.

0 likes 19mo
I
IMainwaringTL3Regular13 Dec 2024 · edited#66

Taking post #65 at face value and following it one step further.

Linearity across the working range is a routine demonstration and it constrains how far a result can be extrapolated. A method linear from 80 to 120 per cent of nominal says nothing about a sample at ten per cent.

This has been discussed before and I could not find the thread, so, again.

32 likes 19mo
SM
so.mbekiTL213 Dec 2024#67
h.nwosu, post #63: Coming back to post #61, because the follow-up matters more than the original answer. Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal). Posting it because the silence on this was… Go to post

A stability-indicating method is one demonstrated to resolve the analyte from its degradation products, usually through forced degradation. Calling a method stability-indicating without that work is a claim rather than a property.

It is worth checking rather than assuming, which costs nothing.

16 likes in reply to #63 19mo
OT
osmolal_tableTL1Member13 Dec 2024#68

Specificity is the first question: does the method separate the analyte from everything reasonably expected to be present? A method that has not been challenged with its own degradation products has not answered it.

6 likes 19mo
VM
v.malinowskiTL213 Dec 2024#69

An independent laboratory's method being different from the supplier's is not a discrepancy. It becomes one only when the results differ by more than both methods' demonstrated precision.

Adding a source would improve this post and I do not have one to hand.

25 likes 19mo
VS
vial_slopeTL3Regular14 Dec 2024#70
IMainwaring, post #66: Taking post #65 at face value and following it one step further. Linearity across the working range is a routine demonstration and it constrains how far a result can be extrapolated. A method linear from 80 to 120 per cent of nominal says nothing about a sample at ten per cent. This has been discussed before and I could not find the… Go to post

Documented validation is what separates a number from an opinion expressed numerically. That is the whole reason to ask for the procedure identifier rather than the method name.

12 likes in reply to #66 19mo
ST
s.teixeiraTL214 Dec 2024#71

Picking up post #68: that is the part I would want checked first.

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

That is the version I use. It may not be the version that is correct.

0 likes 19mo
DN
desiccant_notesTL2Member14 Dec 2024#72

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

Adding it in case it saves somebody the afternoon it cost me.

4 likes 19mo
MI
m.ilungaTL214 Dec 2024#73
ambient_draft, post #54: System suitability is the ongoing evidence that a validated method is still performing. Validation is done once; suitability is done every run, and it is the one that appears on a certificate. Go to post

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

12 likes in reply to #54 19mo
MF
m.ferrandTL1Member14 Dec 2024#74

Robustness: the method gives consistent results when minor parameters vary. Tested by deliberately varying pH, temperature, flow rate, and mobile phase composition within reasonable ranges and demonstrating that results stay within acceptance.

26 likes 19mo
AL
a.lindqvistTL215 Dec 2024#75

This follows post #72 rather than contradicting it.

The distinction between a qualified instrument and a validated method is worth keeping. Both are needed and they fail in different ways.

Happy to be corrected if someone holds better data than mine.

1 like 19mo
JV
j.vandermolenTL3Regular15 Dec 2024#76

Worth separating two things that post #74 runs together.

A method that reports the same number for every lot is worth a second look. Real processes vary, and a total absence of variation is a statement about the measurement rather than the process.

That is the practical version. The rigorous version is longer and says the same thing.

7 likes 19mo
SR
s.roosTL215 Dec 2024#77
baseline_drift, post #49: A method that reports the same number for every lot is worth a second look. Real processes vary, and a total absence of variation is a statement about the measurement rather than the process. Go to post

Sample stability during analysis is part of validation and is routinely ignored. A preparation that degrades in the autosampler over eight hours produces a sequence-dependent result.

I would hold that lightly until someone with a larger sample weighs in.

18 likes in reply to #49 19mo
GV
g.valckenaereTL3Regular15 Dec 2024 · edited#78

That is clearer than the version I had in my head. Thank you.

0 likes 19mo
FR
f.rasmussenTL215 Dec 2024#79

Following this. I have the same question and no better information than the first post.

0 likes 19mo
PR
policy_readerTL2Regular16 Dec 2024#80

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

0 likes 19mo
TS
taper_shiftTL3Regular16 Dec 2024#81

The most useful single question about a method: what would it fail to detect? Every method has an answer and few documents state it.

The short answer was in the first line; everything after is the working.

5 likes 19mo
WM
w.moreauTL216 Dec 2024#82

Reading rather than answering, but this is the post I would point somebody at.

0 likes 19mo
F
FFaulknerTL3Regular16 Dec 2024#83

Answering the question post #80 raises rather than the one it answers.

Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating.

29 likes 19mo
SM
so.mbekiTL216 Dec 2024#84
j.hartmann, post #45: Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples. Scoping that to what I have actually seen rather than what I have read. Go to post

Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance.

14 likes in reply to #45 19mo
MS
m.stephanopoulosTL3Regular17 Dec 2024#85

The limit of quantitation determines what the impurity table can honestly contain. Peaks below it can be reported as detected and cannot be reported as a number.

I am describing what is, rather than arguing for what should be.

9 likes 19mo
JS
j.silvaTL217 Dec 2024#86

Taking post #83 at face value and following it one step further.

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

This is where my knowledge stops and I would rather mark the edge than blur it.

2 likes 19mo
O
OstrowskiTL2Member17 Dec 2024#87

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

0 likes 19mo
HN
h.nwosuTL217 Dec 2024#88
l.sarkissian, post #56: Everything in post #53 holds. The case it does not cover is the one I have. Reporting a result to more decimal places than the method's precision supports is a small dishonesty that appears everywhere. A method with a two per cent relative standard deviation does not support a figure quoted to a hundredth. The honest answer is that it… Go to post

A method transferred between laboratories needs a demonstration that it performs equivalently, not just a document describing it. Transfer is where a great many between-laboratory disagreements originate.

That is all the detail I have. Someone else will have more.

20 likes in reply to #56 19mo
DB
d.bakkerTL217 Dec 2024#89
n.torrence, post #37: Reading rather than contributing, but this is the most useful thread I have found on it. Go to post

I had written a reply contradicting post #87 and deleted it. Here is what survived.

Reporting a result to more decimal places than the method's precision supports is a small dishonesty that appears everywhere. A method with a two per cent relative standard deviation does not support a figure quoted to a hundredth.

That is my reading. Someone else read the same page differently and was reasonable.

13 likes in reply to #37 19mo
AA
a.asanteTL217 Dec 2024#90

Range and working range are different things and a certificate rarely distinguishes them. The relevant one is the range over which this particular sample was measured.

5 likes 19mo