Solved by r.marsden in post #7
Everything in post #3 holds. The case it does not cover is the one I have. Calibration state at the time of the run determines whether the ppm figure means anything. A report that states when the instrument was last calibrated is unusual and is worth more than one that does not.

Jump to the accepted answer →

SV
s.vanheckeTL220 May 2026#1

MS/MS sequence confirmation: what it costs and when it is worth it — setting out what I have, and where I think it stops being reliable.

I would like to understand what this number means before I repeat it anywhere.

A PeptideMeter report on a retatrutide lot gives 98.4% purity. The supplier certificate for the same lot states 99%. Both documents name a reversed-phase method; neither states the same gradient.

My question is not "who is right". It is: given that those two figures were produced by different methods, what is the largest difference I should expect from method alone, and at what point does a gap stop being explainable that way?

24 likes 2mo
SL
s.lindqvistTL26 Jun 2026#2

This follows the opening post rather than contradicting it.

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

4 likes 2mo
ST
sterile_tableTL3Regular18 Jun 2026#3

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

Reporting the observation and leaving the explanation open deliberately.

0 likes 1mo
GB
g.bakkenTL228 Jun 2026#4
s.vanhecke, post #1: MS/MS sequence confirmation: what it costs and when it is worth it — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I repeat it anywhere. A PeptideMeter report on a retatrutide lot gives 98.4% purity. The supplier certificate for the same lot states 99%. Both… Go to post

Isotope pattern is a check that costs nothing to look at. A species at this molecular weight has a characteristic envelope, and something with the wrong envelope is not what it says it is.

A qualification I should have led with rather than closed on.

0 likes in reply to #1 29d
LM
lyophil_marginTL3Regular8 Jul 2026#5
g.bakken, post #4: Isotope pattern is a check that costs nothing to look at. A species at this molecular weight has a characteristic envelope, and something with the wrong envelope is not what it says it is. A qualification I should have led with rather than closed on. Go to post

Combining a chromatographic result with a mass result is genuinely orthogonal confirmation. Either alone leaves a specific class of problem invisible, and the two classes do not overlap much.

Noting that I have skin in this question and have tried to discount for it.

8 likes in reply to #4 19d
AA
a.amankwahTL218 Jul 2026#6
RM
r.marsdenTL3Regular Solution27 Jul 2026#7

Everything in post #3 holds. The case it does not cover is the one I have.

Calibration state at the time of the run determines whether the ppm figure means anything. A report that states when the instrument was last calibrated is unusual and is worth more than one that does not.

7 likes 1d

Suggested topics

TopicParticipantsRepliesViewsActivity
Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a…
JPERBEBWBJ+124 146 15k 16mo
Counter-ion content and its effect on measured mass
On the subject in the title: Counter-ion content and its effect on measured mass Working notes rather than a conclusion. Something about counter-ion content does not reconcile and I would like a second pair…
JTFAIDGHIO+100 108 45k 12mo
Mass error in ppm: the arithmetic and the caveat
Mass error in ppm: the arithmetic and the caveat Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
KCCIAMCDBV+108 125 1.5k 5mo
Distinguishing a deletion sequence by mass alone
Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Trying to establish the boundary conditions on distinguishing a deletion…
RMLSMHMINM+121 129 25k 17mo
What a 2 ppm match establishes, and what it does not
What a 2 ppm match establishes, and what it does not — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
JVBPSOOAK+56 60 5.2k 9mo

Related topics — sharing the tags LC-MS, MALDI-TOF, charge states

TopicParticipantsRepliesViewsActivity
Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below. I have spent a fortnight trying to pin mass spectrometry down…
DAAPINOSS+75 82 4k 17mo
Adducts: sodium, potassium, and the peak you did not expect
Adducts: sodium, potassium, and the peak you did not expect Writing it up because I had to work it out twice and would rather nobody else did. Adducts keeps being re-asked here and I think that is because the…
HMDHNCJWSV+70 77 3.5k 15mo
Isotope patterns and why the monoisotopic peak is not the tallest
Isotope patterns and why the monoisotopic peak is not the tallest — setting out what I have, and where I think it stops being reliable. Isotope patterns: what I expected, what I found, and the gap between…
EKFHPNMBUC+12 16 7.5k 17mo
Second pass at: MALDI versus ESI for a quick identity check
Second pass at: MALDI versus ESI for a quick identity check Writing it up because I had to work it out twice and would rather nobody else did. Something about MALDI versus ESI does not reconcile and I would…
MALKO 2 18k 7mo
What a 2 ppm match establishes, and what it does not
What a 2 ppm match establishes, and what it does not — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
JVBPSOOAK+56 60 5.2k 9mo