Where I would push back on the charge states consensus is the confidence, not the direction. The direction looks right. The confidence is borrowed.
Charge states for a 4 kDa peptide, worked through
Coming back to post #62, because the follow-up matters more than the original answer.
Charge states is worth one more sentence than it usually gets, and the sentence is the one about how the number was arrived at.
Building on post #69 rather than restating it.
An acylated peptide has a mass that reflects the modification, so comparing against the mass of the unmodified backbone gives a mismatch that is not an error.
On reflection I would soften that slightly.
Charge states observed: for semaglutide (4113.6 Da) the doubly charged ion appears at m/z ≈ 2057, triply charged at ≈ 1371, quadruply at ≈ 1029. Those are the positions to look for; the heights depend on the ionization efficiency.
Scoping that to what I have actually seen rather than what I have read.
Right, and stated more narrowly than I would have dared to state it.
Read the full topic (147 posts)
This topic was referenced in
- Distinguishing a deletion sequence by mass alone — the long versionAnalytics › Mass spectrometry · 44 replies
Suggested topics
| Topic | Participants | Replies | Views | Activity |
|---|---|---|---|---|
|
Reading an MS report that only gives you a single number — the long version
Reading an MS report that only gives you a single number — the long version — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I…
|
2 | 47k | 18mo | |
|
Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below. I have spent a fortnight trying to pin mass spectrometry down…
|
+75 | 82 | 4k | 17mo |
|
Isotope patterns and why the monoisotopic peak is not the tallest
Isotope patterns and why the monoisotopic peak is not the tallest — setting out what I have, and where I think it stops being reliable. Isotope patterns: what I expected, what I found, and the gap between…
|
+12 | 16 | 7.5k | 17mo |
|
MALDI versus ESI for a quick identity check
On the subject in the title: MALDI versus ESI for a quick identity check Working notes rather than a conclusion. I was wrong about MALDI versus ESI in a thread last spring and I would like to correct it…
|
2 | 2k | 11mo | |
|
Mass error in ppm: the arithmetic and the caveat — a second dataset
On the subject in the title: Mass error in ppm: the arithmetic and the caveat — a second dataset Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked.…
|
+20 | 24 | 1.7k | 1d |
Related topics — sharing the tags high-resolution MS, counter-ion content, charge states
| Topic | Participants | Replies | Views | Activity |
|---|---|---|---|---|
|
Isotope patterns and why the monoisotopic peak is not the tallest — does this still hold?
Isotope patterns and why the monoisotopic peak is not the tallest — does this still hold? — that is the question, and I have not found it answered plainly anywhere I have looked. Asking about Isotope patterns…
|
2 | 4.6k | 7h | |
|
Adsorption to container surfaces and to plastics
On the subject in the title: Adsorption to container surfaces and to plastics Working notes rather than a conclusion. I was wrong about adsorption in a thread last spring and I would like to correct it…
|
+13 | 17 | 4.3k | 5mo |
|
Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below. I have spent a fortnight trying to pin mass spectrometry down…
|
+75 | 82 | 4k | 17mo |
|
Oxidation and deamidation mass shifts, tabulated
Posting this under the heading it deserves: Oxidation and deamidation mass shifts, tabulated Everything below is what sits behind that. Oxidation and deamidation mass shifts, and specifically the version of…
|
+67 | 76 | 21k | 16mo |
|
About the Mass spectrometry category
LC-MS and high-resolution MS, charge states, mass accuracy in ppm, and what identity confirmation does and does not establish. This post is a community wiki: any member at trust level 3 or above can edit it,…
|
+5 | 9 | 19k | 6h |