Distinguishing a deletion sequence by mass alone — the long version posts 31–45

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1.

IS
isotonic_sheetTL3Regular4 Jul 2025#31

Where I part company with post #29, and it is a narrow parting.

Sample preparation for mass spectrometry can itself introduce modifications, particularly oxidation. A finding at trace level may be telling you about the preparation rather than the material.

0 likes 13mo
PK
p.krastevTL28 Jul 2025#32

Seconded. It reads as careful rather than confident, which is the right register.

25 likes 13mo
RV
r.venkatesanTL3Wiki editor13 Jul 2025#33
e.silva, post #6: Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size. The honest answer is that it depends, and here is what it depends on. Go to post

Mass error in parts per million is (observed minus theoretical) divided by theoretical, times a million. On a high-resolution instrument a low single-digit figure is unremarkable and expected.

It is the sort of thing that seems obvious in retrospect and was not at the time.

12 likes in reply to #6 13mo
FD
f.danquahTL217 Jul 2025 · edited#34

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

The evidence for this is thinner than the way I have phrased it suggests.

4 likes 12mo
TK
t.kulkarniTL321 Jul 2025#35
AV
a.vermeulenTL226 Jul 2025#36

Building on post #33 rather than restating it.

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

0 likes 12mo
RJ
r.jhannsdttirTL3Regular30 Jul 2025#37
p.frisk, post #3: Building on post #2 rather than restating it. Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). Go to post

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

A modest claim, modestly supported.

17 likes in reply to #3 12mo
NK
ni.kravchenkoTL23 Aug 2025#38
isotonic_sheet, post #31: Where I part company with post #29, and it is a narrow parting. Sample preparation for mass spectrometry can itself introduce modifications, particularly oxidation. A finding at trace level may be telling you about the preparation rather than the material. Go to post

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

Written from notes rather than memory, which is why the numbers are specific.

7 likes in reply to #31 12mo
CL
coldchain_liuTL3Regular8 Aug 2025#39

Response in electrospray is not proportional to abundance across different species. Using peak intensities from a mass spectrum to estimate proportions is a mistake that looks reasonable.

The interesting part of this is the exception, and I do not understand the exception.

1 like 12mo
MN
m.nascimentoTL212 Aug 2025#40

Monoisotopic and average mass are different numbers and both appear on documents. At this molecular weight the difference is a few daltons, which is more than the tolerance being claimed.

Worth one more sentence than it usually gets.

0 likes 12mo
CP
citation_peakTL316 Aug 2025#41
AC
a.cabreraTL220 Aug 2025 · edited#42
methods_margin, post #15: Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass. Take it as a starting point and not as a specification. Go to post

Tandem mass spectrometry with fragmentation gives sequence information that intact mass cannot. It is the analysis that would actually confirm identity, and it is rarely supplied.

The general case is well covered; this is the awkward specific one.

3 likes in reply to #15 11mo
KR
k.redgraveTL2Member24 Aug 2025#43
n.krastev, post #29: Taking post #28 at face value and following it one step further. An acylated peptide has a mass that reflects the modification, so comparing against the mass of the unmodified backbone gives a mismatch that is not an error. If the premise is wrong, everything after it is decoration. Go to post

Picking up post #42: that is the part I would want checked first.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

If anyone has run this properly I would rather read that than my own guess.

17 likes in reply to #29 11mo
SI
s.ivaturiTL229 Aug 2025#44

On post #40 — agreed on the reasoning, with one qualification.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

Not a strong opinion, just a consistent one.

32 likes 11mo
R
RidgewayTL3Regular2 Sep 2025#45

Adducts — sodium and potassium especially — shift the observed mass in predictable increments. A report showing an unexplained plus twenty-two is usually showing you a sodium adduct.

None of the above is medical advice and I am not qualified to give any.

0 likes 11mo

Suggested topics

TopicParticipantsRepliesViewsActivity
Second pass at: What a 2 ppm match establishes, and what it does not
Second pass at: What a 2 ppm match establishes, and what it does not — setting out what I have, and where I think it stops being reliable. Two things I would like separated before anyone answers on 2 ppm…
BSNKSTMGMM+44 49 8.1k 1d
Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below. I have spent a fortnight trying to pin mass spectrometry down…
DAAPINOSS+75 82 4k 17mo
MALDI versus ESI for a quick identity check
On the subject in the title: MALDI versus ESI for a quick identity check Working notes rather than a conclusion. I was wrong about MALDI versus ESI in a thread last spring and I would like to correct it…
RMTVO 2 2k 11mo
Distinguishing a deletion sequence by mass alone
Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Trying to establish the boundary conditions on distinguishing a deletion…
RMLSMHMINM+121 129 25k 17mo
About the Mass spectrometry category
LC-MS and high-resolution MS, charge states, mass accuracy in ppm, and what identity confirmation does and does not establish. This post is a community wiki: any member at trust level 3 or above can edit it,…
DOOBLDBDRV+5 9 19k 6h

Related topics — sharing the tags high-resolution MS, mass error (ppm), charge states

TopicParticipantsRepliesViewsActivity
About the Mass spectrometry category
LC-MS and high-resolution MS, charge states, mass accuracy in ppm, and what identity confirmation does and does not establish. This post is a community wiki: any member at trust level 3 or above can edit it,…
DOOBLDBDRV+5 9 19k 6h
Mass error in ppm: the arithmetic and the caveat — a second dataset
On the subject in the title: Mass error in ppm: the arithmetic and the caveat — a second dataset Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked.…
HKGVDWONW+20 24 1.7k 1d
[2026 update] Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
GTAPFP 2 1k 8mo
Isotope patterns and why the monoisotopic peak is not the tallest — does this still hold?
Isotope patterns and why the monoisotopic peak is not the tallest — does this still hold? — that is the question, and I have not found it answered plainly anywhere I have looked. Asking about Isotope patterns…
PODBTD 2 4.6k 7h
Coming back to: Deconvolution of a multiply charged envelope
On the subject in the title: Deconvolution of a multiply charged envelope Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. tirzepatide has a…
VRCVEL 2 12k 22mo