[2026 update] Why mass spectrometry cannot tell you about an epimer

GT
g.tanakaTL3Regular22 Nov 2025#1

Why mass spectrometry cannot tell you about an epimer — that is the question, and I have not found it answered plainly anywhere I have looked.

Working through the identity arithmetic and I would like it checked.

retatrutide has a monoisotopic mass close to 4731.3 Da. On an electrospray instrument I would expect to see the multiply charged series rather than the intact singly charged ion, so for the doubly charged species I calculate (4731.3 + 2 x 1.00728) / 2, and for the triply charged the analogous expression.

The observed values in the report sit within a few ppm of those. My question is what that actually establishes, because I have seen people treat a mass match as a purity result and I do not think it is one.

9 likes 8mo
AP
a.pereiraTL22 Dec 2025#2

Where I part company with the opening post, and it is a narrow parting.

The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.

If it helps: the failure mode here is usually boring rather than dramatic.

13 likes 8mo
FP
forest_plotTL3Evidence synthesis8 Dec 2025#3
a.pereira, post #2: Where I part company with the opening post, and it is a narrow parting. The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had. If it helps: the failure mode here is usually boring rather than dramatic. Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

0 likes in reply to #2 8mo
Moved from Home & field testing by bench_notes. Category placement is not obvious from outside and getting it wrong is expected. This topic will get better answers here. The move is recorded in the public log citing R7.

Suggested topics

TopicParticipantsRepliesViewsActivity
Distinguishing a deletion sequence by mass alone
Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Trying to establish the boundary conditions on distinguishing a deletion…
RMLSMHMINM+121 129 25k 17mo
Reading an MS report that only gives you a single number — the long version
Reading an MS report that only gives you a single number — the long version — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I…
TTRAK 2 47k 18mo
High-resolution MS on aggregates and what it can see — one year on
High-resolution MS on aggregates and what it can see — one year on Writing it up because I had to work it out twice and would rather nobody else did. A narrow question about High-resolution, deliberately…
NGNDBEIWCO+54 58 1.2k 23mo
What a 2 ppm match establishes, and what it does not
What a 2 ppm match establishes, and what it does not — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
JVBPSOOAK+56 60 5.2k 9mo
Coming back to: Oxidation and deamidation mass shifts, tabulated
Oxidation and deamidation mass shifts, tabulated — setting out what I have, and where I think it stops being reliable. Reading back through what has been written here about Oxidation and deamidation mass…
HFSLADAKTI+34 38 4.8k 10mo

Related topics — sharing the tags high-resolution MS, counter-ion content, MALDI-TOF

TopicParticipantsRepliesViewsActivity
Deletion sequences: expected mass shifts by residue
Deletion sequences: expected mass shifts by residue — setting out what I have, and where I think it stops being reliable. Deletion sequences keeps being re-asked here and I think that is because the answer is…
SCBBCN 2 59 3mo
Endotoxin and why it is a separate question from purity — what changed since
On the subject in the title: Endotoxin and why it is a separate question from purity — what changed since Working notes rather than a conclusion. What changes if the standard account of Endotoxin is wrong? I…
JSPRMVMDEM+78 91 1k 10h
Preservatives: function, testing, and limits
Preservatives: function, testing, and limits — setting out what I have, and where I think it stops being reliable. An honest uncertainty about preservatives rather than a disguised assertion. I do not know…
JDMAVSNHST+23 27 3.7k 16mo
Adsorption to container surfaces and to plastics
On the subject in the title: Adsorption to container surfaces and to plastics Working notes rather than a conclusion. I was wrong about adsorption in a thread last spring and I would like to correct it…
CTSOCWRMML+13 17 4.3k 5mo
Deamidation and the close-eluting pair it produces — the long version
Posting this under the heading it deserves: Deamidation and the close-eluting pair it produces — the long version Everything below is what sits behind that. A narrow question about Deamidation, deliberately…
KMMNOAHR 3 347 24d