B
BGiordanoTL2Member31 May 2026#31
s.teixeira, post #17: Confirming post #16 from a second method, which matters more than confirming it from a second person. An itemised related-substances table with retention times says considerably more about the synthesis than a single total. A total tells you how much is not the main peak and nothing about what it is. The strength of my opinion here… Go to post

The distinction between release testing and characterisation is one industrial documents make and retail ones usually do not. It is the difference between "we tested every lot for this" and "we established this once".

I have deliberately not rounded that, because the rounding is where the argument starts.

7 likes in reply to #17 2mo
BV
b.vestergaardTL22 Jun 2026#32

Counter-ion form is crucial: peptides are isolated as salts, most often trifluoroacetate from reversed-phase purification or acetate after salt exchange. The counter-ion is part of the mass in the vial and is not part of the peptide. Two vials of the same purity, one as TFA and one as acetate, contain different amounts of peptide per milligram.

That is the version I would defend. It is not the version I started with.

18 likes 2mo
MC
m.coelhoTL23 Jun 2026#33

Taking post #30 at face value and following it one step further.

Reading old certificates: if a certificate is dated years ago, the lot it describes might be old. That is not automatically a problem but it is worth noting. Fresh testing on the current lot is stronger than old testing on an older lot.

0 likes 2mo
BS
b.solbergTL25 Jun 2026#34
HM
h.mbekiTL26 Jun 2026#35

A certificate is a statement by its issuer that a defined test was performed on a defined lot and produced a defined result. That is all it is, and it is not nothing.

The rule of thumb is fine; the edge cases are where it earns its keep.

4 likes 2mo
KR
k.radichTL28 Jun 2026#36

When method identifiers are missing: "reversed-phase HPLC" is less specific than "reversed-phase HPLC at 214 nm on a C18 column with a 10 to 40% acetonitrile gradient over 20 minutes". The second lets someone reproduce the analysis. The first does not.

12 likes 2mo
BO
b.oseiTL29 Jun 2026#37

This follows post #35 rather than contradicting it.

An itemised related-substances table with retention times says considerably more about the synthesis than a single total. A total tells you how much is not the main peak and nothing about what it is.

I have separated what I observed from what I concluded, which does not always happen.

0 likes 2mo
AR
a.reyesTL4 Admin11 Jun 2026#38
a.lindqvist, post #13: Endotoxin testing is separate: a certificate of purity says nothing about endotoxin. A chemically pure preparation can carry a clinically significant endotoxin load. If endotoxin matters, it needs to be tested and reported. That distinction has done more work for me than anything else in this category. Go to post

A chromatogram supplied as a small image is legible for peak shape and not for baseline detail. That is enough to sanity-check an integration and not enough to reproduce it.

I would treat the number as indicative rather than as a measurement.

0 likes in reply to #13 2mo
LS
l.salinasTL212 Jun 2026#39
a.thorne, post #30: A chromatogram supplied as a small image is legible for peak shape and not for baseline detail. That is enough to sanity-check an integration and not enough to reproduce it. Go to post

Appreciated. The plain phrasing does more work here than a longer post would.

2 likes in reply to #30 2mo
SL
s.leclercTL4 Moderator14 Jun 2026#40
l.salinas, post #39: Appreciated. The plain phrasing does more work here than a longer post would. Go to post

Post #36 describes the usual case. This is about the unusual one.

The distinction between release testing and characterisation is one industrial documents make and retail ones usually do not. It is the difference between "we tested every lot for this" and "we established this once".

8 likes in reply to #39 1mo
NT
nl_translatorTL2Translator · NL15 Jun 2026#41

I read post #37 twice before replying, because I had assumed the opposite.

A document that puts the specification in the result column has told you the release criterion rather than the measurement. Those are different claims and the difference is the whole point of reading the document.

A partial answer, offered because a partial answer beats none.

4 likes 1mo
BD
b.dumitruTL217 Jun 2026 · edited#42
b.solberg, post #34: Helpful, and easy to find again, which is half of what a good reply is. Go to post

Post #41 answers the question as asked. The question underneath it is different.

"Not less than 98 per cent" is a specification. "98.3 per cent" is a result. A certificate carrying only the first has not told you what was measured.

0 likes in reply to #34 1mo
DB
d.bramleyTL3Regular18 Jun 2026#43

A chromatogram supplied as a small image is legible for peak shape and not for baseline detail. That is enough to sanity-check an integration and not enough to reproduce it.

0 likes 1mo
VB
v.bruunTL220 Jun 2026#44

A certificate of analysis is a statement by its issuer that a defined test was performed on a defined lot and produced a defined result. That is all it is, and it is not nothing. What makes it usable: lot identifier matching the container, test date, method identification specific enough to reproduce, the actual result as a number, and the acceptance criterion stated separately from the result.

19 likes 1mo
DM
d.moreauTL2Regular21 Jun 2026#45

Coming back to post #41, because the follow-up matters more than the original answer.

Endotoxin testing is separate: a certificate of purity says nothing about endotoxin. A chemically pure preparation can carry a clinically significant endotoxin load. If endotoxin matters, it needs to be tested and reported.

8 likes 1mo
RL
r.lundgrenTL223 Jun 2026#46
b.dumitru, post #42: Post #41 answers the question as asked. The question underneath it is different. "Not less than 98 per cent" is a specification. "98.3 per cent" is a result. A certificate carrying only the first has not told you what was measured. Go to post

What fields matter on a certificate: lot number matching the vial, test date, the analytical method stated specifically, the measured result as a number, and the acceptance limit stated separately. A certificate missing any of these is weaker.

2 likes in reply to #42 1mo
EN
electrolyte_notesTL2Regular24 Jun 2026#47

An itemised related-substances table with retention times says considerably more about the synthesis than a single total. A total tells you how much is not the main peak and nothing about what it is.

I would put a moderate confidence on that and no more.

0 likes 1mo

Suggested topics

TopicParticipantsRepliesViewsActivity
Second pass at: A field-by-field walk through a certificate of analysis
Second pass at: A field-by-field walk through a certificate of analysis Writing it up because I had to work it out twice and would rather nobody else did. The question about field-by-field walk through a…
ANTMPCNEA+149 164 36k 19mo
[2026 update] A certificate with no method stated: what can you conclude?
A certificate with no method stated: what can you conclude? — that is the question, and I have not found it answered plainly anywhere I have looked. I would like to understand what this number means before I…
IIRV 2 82 11mo
Water content on a certificate and why it changes the arithmetic — what changed since
On the subject in the title: Water content on a certificate and why it changes the arithmetic — what changed since Working notes rather than a conclusion. What changes if the standard account of Water content…
VKIDTBBP 3 801 8mo
When a tidy document is a signal and when it is decoration — does this still hold?
Asking directly, because I could not find a straight answer: When a tidy document is a signal and when it is decoration — does this still hold? Tidy document keeps being re-asked here and I think that is…
VAFKTDJN+3 7 22k 11mo
A field-by-field walk through a certificate of analysis
Posting this under the heading it deserves: A field-by-field walk through a certificate of analysis Everything below is what sits behind that. I have spent a fortnight trying to pin field-by-field walk…
ARVMNKKHN+136 147 1.8k 12mo

Related topics — sharing the tags certificate of analysis, batch & lot numbers, amino acid analysis

TopicParticipantsRepliesViewsActivity
Certificates issued by the seller versus by a laboratory
Posting this under the heading it deserves: Certificates issued by the seller versus by a laboratory Everything below is what sits behind that. I would like to understand what this number means before I…
GACTDEFJV+58 65 1.2k 15mo
Revisiting: A field-by-field walk through a certificate of analysis
Revisiting: A field-by-field walk through a certificate of analysis Writing it up because I had to work it out twice and would rather nobody else did. I would like to disagree carefully with the settled view…
HFRJPNEAPK+37 41 656 3mo
About the COA interpretation category
What a certificate of analysis commits its issuer to, which fields matter, and how to spot one that says almost nothing. This post is a community wiki: any member at trust level 3 or above can edit it, and…
JMHSHVSKLF+8 12 8.5k 2mo
The report format the maintainers ask for, and why each field is there
On the subject in the title: The report format the maintainers ask for, and why each field is there Working notes rather than a conclusion. Structured report rather than an opinion, following the format the…
CHKKCWRBTN 4 6.3k 1d
Follow-up: Reconciling a label claim with a content assay
Reconciling a label claim with a content assay Writing it up because I had to work it out twice and would rather nobody else did. A follow-up question about Reconciling a label claim that I did not know to…