Post #7 is right about the mechanism and I think understates the practical bit.
A chromatogram image at a resolution where you can see peak shape but not baseline detail is worth having and is not the same as the data. Ask for the integration table if the number matters.
22 likes 10mo
n.moreau, post #2: Where I part company with the opening post, and it is a narrow parting. Where the baseline is drawn is the least documented and most consequential choice in the whole determination. Valley-to-valley, tangent skim and forced-to-zero can span a couple of percentage points on the same chromatogram. Adding a source would improve this post… Go to post
A relative retention time against a known peak travels much better than an absolute one, and almost nobody reports it.
The general case is well covered; this is the awkward specific one.
a.frisk, post #4: What I would check first on integrator is whether the thing being measured moved or whether the way of measuring it moved. Those look identical in a graph. Go to post
Adding a note of thanks rather than an opinion. I did not know most of that.
What I can speak to on integrator is narrow, so I will keep it narrow rather than generalising from it. Beyond that boundary I do not know.
20 likes 10mo
Read the full topic (25 posts)
This topic was referenced in
- Why 99.2% and 97.8% on the same lot can both be correctAnalytics › HPLC & UHPLC · 20 replies
Moved from COA interpretation by KLindqvist. Category placement is not obvious from outside and getting it wrong is expected. This topic will get better answers here. The move is recorded in the public log citing R7.
Suggested topics
| Topic | Participants | Replies | Views | Activity |
|---|---|---|---|---|
|
Gradient steepness and apparent purity: a worked comparison — the long version
Gradient steepness and apparent purity: a worked comparison — the long version — setting out what I have, and where I think it stops being reliable. A narrow question about Gradient steepness and apparent…
|
2 | 32k | 1d | |
|
Theoretical plates: useful number or ritual?
Theoretical plates: useful number or ritual? I have a specific reason for asking rather than idle curiosity, and the context is below. Asking about theoretical plates directly, because I have read four…
|
2 | 284 | 9mo | |
|
Detection wavelength and why 214 nm and 280 nm disagree
On the subject in the title: Detection wavelength and why 214 nm and 280 nm disagree Working notes rather than a conclusion. I would like to disagree carefully with the settled view on detection wavelength,…
|
+78 | 84 | 1.7k | 8mo |
|
Column chemistry choices for a 40-residue peptide
Column chemistry choices for a 40-residue peptide — setting out what I have, and where I think it stops being reliable. Collecting what is known about column chemistry choices in one place, because it is…
|
+39 | 43 | 1.1k | 10mo |
|
Gradient steepness and apparent purity: a worked comparison
On the subject in the title: Gradient steepness and apparent purity: a worked comparison Working notes rather than a conclusion. I have spent a fortnight trying to pin gradient steepness and apparent purity…
|
2 | 51k | 22mo |
Related topics — sharing the tags peak integration, gradient method, UV & DAD detection
| Topic | Participants | Replies | Views | Activity |
|---|---|---|---|---|
|
Injection volume, overload, and peak distortion — one year on
Injection volume, overload, and peak distortion — one year on Writing it up because I had to work it out twice and would rather nobody else did. Posting a small dataset on Injection volume. It is mine, it is…
|
+4 | 8 | 1.2k | 4mo |
|
Forced degradation studies: designing one that is informative — does this still hold?
The question in the title: Forced degradation studies: designing one that is informative — does this still hold? I will give what I have already checked below so nobody repeats it. The question about Forced…
|
+37 | 41 | 1.2k | 12mo |
|
Revisiting: Sample handling between despatch and injection
Posting this under the heading it deserves: Revisiting: Sample handling between despatch and injection Everything below is what sits behind that. Reading back through what has been written here about Sample…
|
+5 | 9 | 1.1k | 6mo |
|
Endotoxin and why it is a separate question from purity — what changed since
On the subject in the title: Endotoxin and why it is a separate question from purity — what changed since Working notes rather than a conclusion. What changes if the standard account of Endotoxin is wrong? I…
|
+78 | 91 | 1k | 10h |
|
Where impurities in solid-phase peptide synthesis come from
Where impurities in solid-phase peptide synthesis come from Writing it up because I had to work it out twice and would rather nobody else did. Impurities in solid-phase peptide synthesis, and specifically the…
|
+54 | 59 | 14k | 3mo |