RM
r.marsdenTL3Regular22 Feb 2025#121

Building on post #118 rather than restating it.

For a compound with no reference standard in circulation, an observed mass is much more useful than an assertion of agreement with a theoretical value nobody can check.

It is the sort of thing that seems obvious in retrospect and was not at the time.

20 likes 17mo
AA
a.amankwahTL223 Feb 2025 · edited#122

Post #120 put the caveat in the right place and I want to underline it.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

That has been true for the cases I have seen and I have not seen many.

0 likes 17mo
P
PSundbergTL2Member23 Feb 2025#123
m.agyeman, post #96: Coming back to post #92, because the follow-up matters more than the original answer. Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy. Go to post

Having read the whole distinguishing a deletion sequence thread before replying: the question in the first post has not actually been answered yet, and three of us have answered a nearby one instead.

2 likes in reply to #96 17mo
FF
f.fontaineTL223 Feb 2025#124
r.mensa, post #1: Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Trying to establish the boundary conditions on distinguishing a deletion sequence. The general claim is fine; I want to know where it stops holding, because that is where I am operating. Three edge cases… Go to post

Trying to state the distinguishing a deletion sequence position in a way that someone who disagrees would recognise as fair, because I do not think the version in this thread passes that test.

9 likes in reply to #1 17mo
KF
k.farrugiaTL3Regular23 Feb 2025#125

Sample preparation for mass spectrometry can itself introduce modifications, particularly oxidation. A finding at trace level may be telling you about the preparation rather than the material.

28 likes 17mo
KO
k.okaforTL224 Feb 2025#126

Agreed, and I will stop repeating the version of this I had been repeating.

0 likes 17mo
L
LeitermanTL324 Feb 2025#127
CB
c.balogunTL224 Feb 2025#128
buffer_review, post #115: I had written a reply contradicting post #113 and deleted it. Here is what survived. Adducts — sodium and potassium especially — shift the observed mass in predictable increments. A report showing an unexplained plus twenty-two is usually showing you a sodium adduct. It cost nothing to check and would have cost something not to. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

13 likes in reply to #115 17mo
AR
ambient_reviewTL3Regular24 Feb 2025 · edited#129

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

0 likes 17mo
AP
a.petrovTL224 Feb 2025#130

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes 17mo

Suggested topics

TopicParticipantsRepliesViewsActivity
[2026 update] Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. semaglutide has a…
DBBIATNAL+116 135 31k 3mo
Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a…
JPERBEBWBJ+124 146 15k 16mo
About the Mass spectrometry category
LC-MS and high-resolution MS, charge states, mass accuracy in ppm, and what identity confirmation does and does not establish. This post is a community wiki: any member at trust level 3 or above can edit it,…
DOOBLDBDRV+5 9 19k 6h
Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below. I have spent a fortnight trying to pin mass spectrometry down…
DAAPINOSS+75 82 4k 17mo
Adducts: sodium, potassium, and the peak you did not expect — does this still hold?
Asking directly, because I could not find a straight answer: Adducts: sodium, potassium, and the peak you did not expect — does this still hold? Adducts keeps being re-asked here and I think that is because…
KBZYLVCETV+113 119 1.7k 3mo

Related topics — sharing the tags mass error (ppm), high-resolution MS, LC-MS

TopicParticipantsRepliesViewsActivity
Mass error in ppm: the arithmetic and the caveat
Mass error in ppm: the arithmetic and the caveat Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like it checked.…
KCCIAMCDBV+108 125 1.5k 5mo
Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. retatrutide has a…
JPERBEBWBJ+124 146 15k 16mo
High-resolution MS on aggregates and what it can see
Posting this under the heading it deserves: High-resolution MS on aggregates and what it can see Everything below is what sits behind that. Something about high-resolution does not reconcile and I would like…
CRGISCBPEC+117 123 6.5k 8mo
Oxidation and deamidation mass shifts, tabulated
Posting this under the heading it deserves: Oxidation and deamidation mass shifts, tabulated Everything below is what sits behind that. Oxidation and deamidation mass shifts, and specifically the version of…
AFJVIATNCS+67 76 21k 16mo
[2026 update] Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
GTAPFP 2 1k 8mo