Asking directly, because I could not find a straight answer: Why one laboratory's 98.4% and another's 96.1% may not contradict — one year on
Posting the method first, because I know what the first three replies will otherwise be.
- Column: C18, 3.0 x 150 mm, 2.6 um
- Mobile phase: 0.1% TFA in water / 0.1% TFA in acetonitrile
- Gradient: 9% to 61% organic over 33 minutes
- Detection: 214 nm
- Injection: 18 uL
- Sample: retatrutide, reconstituted to 2.0 mg/mL, injected within an hour
The main peak integrates at 99.1% of total area. There is a small feature on the trailing edge that I cannot decide is a shoulder or a baseline artefact, and that is what I am actually asking about.