Speaking only to Carryover as I have actually seen it, rather than as it is usually described: the effect is real, it is smaller than the thread suggests, and the variance between people is larger than the effect.

0 likes 28d
PM
physio_marchettiTL2Physiotherapist1 Jul 2026#92

Peaks that do not elute do not appear in the area percentage. Aggregates and strongly retained species can be entirely invisible to a standard method, which is a ceiling on what any purity figure can claim.

It is one reading of the data and not the only reasonable one.

0 likes 27d
JI
j.iyerTL21 Jul 2026#93

Taking post #90 at face value and following it one step further.

Before anything else: what was the gradient, and at what wavelength? Area percent at different wavelengths is not the same number even on the same sample because different species absorb differently at different wavelengths. With the method stated, I can tell you something useful. Without it, all I can say is that there is one large peak.

7 likes 27d
CL
coldchain_liuTL3Regular1 Jul 2026#94
m.radich, post #70: Change the wavelength and the proportions change even though the sample has not. That is the reason the wavelength has to be on the document for the number to mean anything. Go to post

Post #92 and I disagree about the size of the effect, not about the direction.

A request rather than an answer: could whoever has the primary source for Carryover post it? I have seen the claim three times this month and each version had lost a qualifier.

18 likes in reply to #70 27d
SK
s.kuuselaTL21 Jul 2026#95
n.abernathy, post #18: Everything in post #16 holds. The case it does not cover is the one I have. Carryover from a previous injection shows up as a small peak at the same retention time in a blank. A method report that includes a blank injection is telling you the analyst checked. Worth saying I have only my own numbers here, and n is small. Go to post

Peak purity: a diode-array detector records a spectrum at every time point. If a peak contains two co-eluting species with different spectra, the spectrum changes across the peak. A passing peak-purity result says the spectrum is constant; it is weak evidence of homogeneity if the impurities have similar spectra.

Someone will know this better than I do and I hope they say so.

26 likes in reply to #18 27d
TY
two_year_lineTL3Regular2 Jul 2026 · edited#96

Method validation is the demonstration that a method can separate the compound from its degradation products and impurities reliably. A method that cannot resolve an impurity from the parent peak will not detect that impurity.

0 likes 26d
AP
au.pereiraTL22 Jul 2026#97
MM
maintenance_modeTL3Regular2 Jul 2026#98

Carryover is one of those subjects where the general answer and the answer for a specific case diverge, and the thread will go in circles until someone says which one is being asked for.

12 likes 26d
KP
k.pereiraTL22 Jul 2026#99

Adding the measurement that post #98 says would settle it.

Column temperature affects retention and selectivity and is omitted from most certificates. Two runs at twenty-five and forty degrees are not the same method.

0 likes 26d
RV
r.venkatesanTL3Wiki editor3 Jul 2026#100

This is the first time the answer has come with its own limits attached. Appreciated.

4 likes 25d
SC
s.cabreraTL23 Jul 2026#101
week_three, post #65: Everything in post #62 holds. The case it does not cover is the one I have. This is why a purity figure without the underlying chromatogram is weaker evidence than it appears. It is also why two competent laboratories can report different numbers on the same vial without either being wrong. Not a strong opinion, just a consistent one. Go to post

Picking up post #99: that is the part I would want checked first.

Typical suitability criteria are a replicate area relative standard deviation below about two per cent, a tailing factor inside a defined window, a resolution minimum against a specified peak, and a plate-count floor.

3 likes in reply to #65 25d
PW
PharmNotes_WhitfieldTL4Pharmacist3 Jul 2026#102

Injection volume matters because column overload distorts peak shape, and an overloaded main peak can swallow a small neighbour. A certificate without injection volume is missing something load-bearing.

If it helps: the failure mode here is usually boring rather than dramatic.

11 likes 25d
NK
n.kuuselaTL23 Jul 2026#103

For anyone finding this later: the short answer on Carryover is that it depends on one thing, and the rest of the thread is people identifying which thing.

25 likes 25d
NA
n.abernathyTL3Analytical chemist3 Jul 2026 · edited#104

I keep a log for Carryover specifically because my memory of it turned out to be systematically wrong in one direction. Six weeks of notes cost nothing and settled it.

0 likes 24d
PM
p.mwangiTL24 Jul 2026#105
n.rahimi, post #79: Practical experience of Carryover, offered as one case with the conditions stated, not as a general finding. Conditions first, because they are what make it interpretable. Go to post

Gradient slope is the single biggest driver of apparent purity differences. A shallower gradient over a longer run resolves more impurities and gives a higher purity figure. A steep gradient produces a tidier-looking chromatogram with fewer visible peaks and gives a lower purity figure. Both are legitimate methods and they will not produce the same number.

On reflection I would soften that slightly.

1 like in reply to #79 24d
DH
dietitian_hollisTL3Dietitian4 Jul 2026#106

The confident answers on Carryover and the well-sourced answers are not the same answers, which is the most useful thing I have learned reading this category.

7 likes 24d
EH
e.halonenTL24 Jul 2026#107

Where I have landed on Carryover, having got it wrong once in public: the direction is clear, the magnitude is not, and anyone quoting a precise magnitude has borrowed it from somewhere that did not measure it.

18 likes 24d
JW
journalclub_wrenTL3Regular4 Jul 2026#108

Where I part company with post #106, and it is a narrow parting.

A relative retention time against a known peak travels much better than an absolute one, and almost nobody reports it.

0 likes 23d
ZL
z.laurentTL25 Jul 2026#109
j.iyer, post #93: Taking post #90 at face value and following it one step further. Before anything else: what was the gradient, and at what wavelength? Area percent at different wavelengths is not the same number even on the same sample because different species absorb differently at different wavelengths. With the method stated, I can tell you something… Go to post

Carryover has been discussed here with more heat than it deserves, mostly because two definitions have been in play the whole time.

11 likes in reply to #93 23d
HE
h.eriksenTL25 Jul 2026#110
Bramley, post #9: Adding the boring version of Carryover, because the interesting version keeps getting posted and the boring one is usually right. Check the ordinary explanations, in order, and stop when one of them accounts for what you are seeing. Most of the time the second one does. Go to post

The most useful thing anyone has posted about Carryover in this category was a table of what had been measured and by whom. That is what I would want again.

24 likes in reply to #9 23d
EN
e.nilsenTL25 Jul 2026#111
f.danquah, post #1: Carryover and the ghost peak from last week's standard — what changed since Writing it up because I had to work it out twice and would rather nobody else did. What changes if the standard account of Carryover is wrong? I ask because I have been treating it as settled and I noticed this week that I could not say why. Working through the… Go to post

Carryover from a previous injection shows up as a small peak at the same retention time in a blank. A method report that includes a blank injection is telling you the analyst checked.

1 like in reply to #1 23d
FT
fr.translation_moTL2Translator · FR5 Jul 2026 · edited#112

Confirming post #109 from a second method, which matters more than confirming it from a second person.

A chromatogram image at a resolution where you can see peak shape but not baseline detail is worth having and is not the same as the data. Ask for the integration table if the number matters.

0 likes 22d
AT
a.teixeiraTL26 Jul 2026#113

On post #109 — agreed on the reasoning, with one qualification.

I have no financial interest in anything named in this thread and I want to say so before I comment on Carryover, because it is the sort of subject where it matters.

23 likes 22d
RF
resistance_firstTL2Regular6 Jul 2026#114

The version of Carryover that circulates here is a simplification of a simplification. It is not wrong, but it has lost the conditions under which it holds, and those conditions are where the interesting cases live.

10 likes 22d
AV
a.vestergaardTL26 Jul 2026#115
n.kuusela, post #103: For anyone finding this later: the short answer on Carryover is that it depends on one thing, and the rest of the thread is people identifying which thing. Go to post

A shallow gradient resolves close-eluting species that a steep one merges. Two honest laboratories running different gradients can report genuinely different purities on the same vial.

It is worth checking rather than assuming, which costs nothing.

0 likes in reply to #103 22d

Suggested topics

TopicParticipantsRepliesViewsActivity
[2026 update] Detection wavelength and why 214 nm and 280 nm disagree
Detection wavelength and why 214 nm and 280 nm disagree — setting out what I have, and where I think it stops being reliable. I would like to disagree carefully with the settled view on Detection wavelength,…
HVKAMCDKR+69 76 920 6h
Why 99.2% and 97.8% on the same lot can both be correct
Why 99.2% and 97.8% on the same lot can both be correct I have a specific reason for asking rather than idle curiosity, and the context is below. I would like to understand what this number means before I…
GHAAJSLCCR+16 20 3.1k 11h
Second pass at: What a reversed-phase purity number actually is
On the subject in the title: Second pass at: What a reversed-phase purity number actually is Working notes rather than a conclusion. I would like to understand what this number means before I repeat it…
JTSPNS 2 7.4k 11mo
Comparing two chromatograms from different laboratories, properly
Posting this under the heading it deserves: Comparing two chromatograms from different laboratories, properly Everything below is what sits behind that. I would like to understand what this number means…
DBKKKFFLA+35 40 22k 7mo
Coming back to: Column chemistry choices for a 40-residue peptide
On the subject in the title: Column chemistry choices for a 40-residue peptide Working notes rather than a conclusion. Column chemistry choices — I have the observation and I do not trust my interpretation of…
MLFPHB 2 43k 16mo

Related topics — sharing the tags gradient method, data table, HPLC

TopicParticipantsRepliesViewsActivity
Coming back to: Specificity, linearity, accuracy, precision, range, robustness — with real criteria
On the subject in the title: Specificity, linearity, accuracy, precision, range, robustness — with real criteria Working notes rather than a conclusion. Trying to establish the boundary conditions on…
AWHNFJCOT+93 99 1.4k 6d
Comparing my trajectory with the trial curves, carefully — a second dataset
Comparing my trajectory with the trial curves, carefully — a second dataset — setting out what I have, and where I think it stops being reliable. Maintenance rather than loss, which is the part of this that…
OPYMHESTBW+77 81 16k 11h
Validating a method you did not develop — does this still hold?
Validating a method you did not develop — does this still hold? I have a specific reason for asking rather than idle curiosity, and the context is below. Validating a method, and specifically the version of…
AAEFNRRBFE+128 148 47k 16mo
Repeatability versus intermediate precision
Repeatability versus intermediate precision — setting out what I have, and where I think it stops being reliable. Trying to establish the boundary conditions on repeatability versus intermediate precision.…
LSKAILWNO+119 134 34k 13mo
[2026 update] Waist circumference technique and its repeatability
Posting this under the heading it deserves: Waist circumference technique and its repeatability Everything below is what sits behind that. An honest uncertainty about Waist circumference technique rather than…
RZZACNSKG+23 27 5.9k 21h