What a stability-indicating method is and how you demonstrate one — one year on posts 91–113

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1 · go to the accepted answer.

JW
journalclub_wrenTL3Regular10 Dec 2024#91

This follows post #89 rather than contradicting it.

Reporting a result to more decimal places than the method's precision supports is a small dishonesty that appears everywhere. A method with a two per cent relative standard deviation does not support a figure quoted to a hundredth.

One of those cases where knowing the mechanism does not help the decision.

27 likes 20mo
NC
n.cabreraTL211 Dec 2024#92

Worth separating two things that post #88 runs together.

Documented validation is what separates a number from an opinion expressed numerically. That is the whole reason to ask for the procedure identifier rather than the method name.

0 likes 20mo
SS
steady_stateTL3Regular12 Dec 2024 · edited#93

My experience of stability-indicating method contradicts the reply above. I am posting it as a data point rather than as a refutation, because one person's experience is exactly that.

4 likes 20mo
SV
s.vukovicTL214 Dec 2024#94
dexa_twice_yearly, post #29: Where I part company with post #25, and it is a narrow parting. Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements. Go to post

Range and working range are different things and a certificate rarely distinguishes them. The relevant one is the range over which this particular sample was measured.

I would treat that as a working assumption and revisit it.

13 likes in reply to #29 19mo
SB
sharps_binTL2Regular15 Dec 2024#95

Taking post #94 at face value and following it one step further.

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

20 likes 19mo
SO
se.okaforTL216 Dec 2024#96

Understood. Thank you for being specific about the limits of it.

0 likes 19mo
OF
outline_firstTL3Wiki editor18 Dec 2024#97

Linearity across the working range is a routine demonstration and it constrains how far a result can be extrapolated. A method linear from 80 to 120 per cent of nominal says nothing about a sample at ten per cent.

The literature is thinner on this than the confidence in the thread implies.

2 likes 19mo
GA
g.amankwahTL219 Dec 2024#98
priorauth_notes, post #40: Post #38 and I disagree about the size of the effect, not about the direction. A method transferred between laboratories needs a demonstration that it performs equivalently, not just a document describing it. Transfer is where a great many between-laboratory disagreements originate. Go to post

I read post #97 twice before replying, because I had assumed the opposite.

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

I would rather be precise about what I do not know than vague about what I do.

8 likes in reply to #40 19mo
BN
bench_notesTL420 Dec 2024#99
AV
a.vukovicTL222 Dec 2024#100

Validation is compound-specific and matrix-specific. A method validated for one peptide is a starting point for another and not a validated method for it.

The general case is well covered; this is the awkward specific one.

9 likes 19mo
GD
glossary_deskTL3Regular23 Dec 2024#101

Adding the measurement that post #100 says would settle it.

Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements.

0 likes 19mo
AK
an.kirchnerTL224 Dec 2024#102

The practical version of stability-indicating method is three sentences long. The rigorous version is three pages and reaches the same conclusion with the conditions attached.

5 likes 19mo
GC
glossary_checkTL2Member25 Dec 2024#103

Robustness testing deliberately varies the parameters most likely to drift — organic percentage, pH, temperature, flow — and shows the result does not. It is the part of validation that predicts whether a method will transfer.

That is where I would start, not where I would stop.

13 likes 19mo
SP
s.perrinTL227 Dec 2024#104
a.sorensen, post #61: Post #57 describes the usual case. This is about the unusual one. Precision and repeatability: within-run and between-run variability of the method. Acceptance criterion is typically a relative standard deviation of ≤2% for area measurements. Go to post

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

Happy to be the one who is wrong here if it settles the question.

28 likes in reply to #61 19mo
CN
cohort_notesTL2Member28 Dec 2024#105
se.okafor, post #96: Understood. Thank you for being specific about the limits of it. Go to post

Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount.

0 likes in reply to #96 19mo
GD
g.danquahTL229 Dec 2024#106

A method that reports the same number for every lot is worth a second look. Real processes vary, and a total absence of variation is a statement about the measurement rather than the process.

That is what the documentation says. What happens in practice is usually close.

2 likes 19mo
B
BuchholzTL2Member31 Dec 2024#107

System suitability is the ongoing evidence that a validated method is still performing. Validation is done once; suitability is done every run, and it is the one that appears on a certificate.

If that is already documented somewhere, ignore me and link it.

9 likes 19mo
FE
f.espinozaTL21 Jan 2025 · edited#108

Post #106 put the caveat in the right place and I want to underline it.

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

The strength of my opinion here exceeds the strength of my evidence.

20 likes 19mo
TW
t.waldenstrmTL2Member2 Jan 2025#109

Useful. I had the fact and not the reason, which turns out to be the important half.

0 likes 19mo
SR
s.radichTL24 Jan 2025#110

On post #106 — agreed on the reasoning, with one qualification.

The most useful single question about a method: what would it fail to detect? Every method has an answer and few documents state it.

I have deliberately not rounded that, because the rounding is where the argument starts.

0 likes 19mo
VK
v.klausenTL3Regular5 Jan 2025#111
l.aaltonen, post #76: Range: the concentration range over which the method has been validated. Going outside the validated range is going outside the method's demonstrated performance. Go to post

Post #110 describes the usual case. This is about the unusual one.

Documented validation is what separates a number from an opinion expressed numerically. That is the whole reason to ask for the procedure identifier rather than the method name.

0 likes in reply to #76 19mo
YR
y.ramosTL26 Jan 2025#112
s.ferreira, post #52: Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount. One more caveat and then I will stop qualifying: the sample selected itself. Go to post

Adding the measurement that post #110 says would settle it.

Offering a way to settle stability-indicating method rather than another opinion about it. Two measurements, taken the same way, a fortnight apart. If the difference is within the noise, the question was not answerable at this precision.

26 likes in reply to #52 19mo
ID
integrator_draftTL3Regular7 Jan 2025#113

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

Anyone with a larger sample, please post it.

12 likes 19mo

Suggested topics

TopicParticipantsRepliesViewsActivity
Specificity, linearity, accuracy, precision, range, robustness — with real criteria
Specificity, linearity, accuracy, precision, range, robustness — with real criteria Writing it up because I had to work it out twice and would rather nobody else did. Trying to establish the boundary…
FNISMLAWM+44 48 12k 1d
LOD and LOQ: definitions people get wrong
LOD and LOQ: definitions people get wrong — setting out what I have, and where I think it stops being reliable. What changes if the standard account of LOD and LOQ is wrong? I ask because I have been treating…
EISFCN 2 43k 8mo
Coming back to: When to suspect the method rather than the sample
When to suspect the method rather than the sample I have a specific reason for asking rather than idle curiosity, and the context is below. Two reports on the same lot, and I would like help deciding whether…
OVGBFRSLRM+116 135 42k 19mo
Repeatability versus intermediate precision
Repeatability versus intermediate precision — setting out what I have, and where I think it stops being reliable. Trying to establish the boundary conditions on repeatability versus intermediate precision.…
LSKAILWNO+119 134 34k 13mo
Revisiting: Sample handling between despatch and injection
Posting this under the heading it deserves: Revisiting: Sample handling between despatch and injection Everything below is what sits behind that. Reading back through what has been written here about Sample…
DVEAGTEMNG+5 9 1.1k 6mo

Related topics — sharing the tags precision & repeatability, system suitability, stability-indicating method

TopicParticipantsRepliesViewsActivity
Where the integrator drew the baseline, and how much it moved the number
Where the integrator drew the baseline, and how much it moved the number — setting out what I have, and where I think it stops being reliable. Posting the method first, because I know what the first three…
FNNPMAVKTV+20 24 1.6k 8mo
Relative retention time as a better inter-laboratory comparator — does this still hold?
Relative retention time as a better inter-laboratory comparator — does this still hold? I have a specific reason for asking rather than idle curiosity, and the context is below. What changes if the standard…
EARFWNFWOO+11 15 21k 21mo
Coming back to: Specificity, linearity, accuracy, precision, range, robustness — with real criteria
On the subject in the title: Specificity, linearity, accuracy, precision, range, robustness — with real criteria Working notes rather than a conclusion. Trying to establish the boundary conditions on…
AWHNFJCOT+93 99 1.4k 6d
Why one laboratory's 98.4% and another's 96.1% may not contradict — one year on
Asking directly, because I could not find a straight answer: Why one laboratory's 98.4% and another's 96.1% may not contradict — one year on Posting the method first, because I know what the first three…
MSFKJBMRCR+10 14 13k 5mo
Coming back to: What consumer-accessible tests can and cannot determine
Asking directly, because I could not find a straight answer: What consumer-accessible tests can and cannot determine Something about consumer-accessible tests does not reconcile and I would like a second pair…
PKNWVNHK+4 8 25k 22mo