I
IbrahimoviTL2Member26 Aug 2025#31

Narrowing post #30, because the general version has more than one answer.

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

I would treat the number as indicative rather than as a measurement.

31 likes 11mo
ET
e.tammTL230 Aug 2025#32
a.reyes, post #28: Narrowing post #27, because the general version has more than one answer. Trifluoroacetate adducts are common in material purified with TFA and are one reason a mass spectrum from a peptide can look busier than expected. The confident version of this sentence would be wrong, so here is the hedged one. Go to post

Response in electrospray is not proportional to abundance across different species. Using peak intensities from a mass spectrum to estimate proportions is a mistake that looks reasonable.

The rule of thumb is fine; the edge cases are where it earns its keep.

0 likes in reply to #28 11mo
CN
cannula_notesTL2Member3 Sep 2025#33

Oxidation adds sixteen per oxygen and is the most common modification seen in aged material. Its presence in the spectrum is informative about handling as well as synthesis.

6 likes 11mo
BB
b.brandtTL26 Sep 2025#34
M
MSaarinenTL3Regular10 Sep 2025#35
s.leclerc, post #26: Sample preparation for mass spectrometry can itself introduce modifications, particularly oxidation. A finding at trace level may be telling you about the preparation rather than the material. That is one dataset and I would not build a rule on it. Go to post

Tandem mass spectrometry with fragmentation gives sequence information that intact mass cannot. It is the analysis that would actually confirm identity, and it is rarely supplied.

I am aware this is the third time this month I have made this point.

23 likes in reply to #26 11mo
TL
t.lindqvistTL214 Sep 2025#36
MSaarinen, post #35: Tandem mass spectrometry with fragmentation gives sequence information that intact mass cannot. It is the analysis that would actually confirm identity, and it is rarely supplied. I am aware this is the third time this month I have made this point. Go to post

Adducts — sodium and potassium especially — shift the observed mass in predictable increments. A report showing an unexplained plus twenty-two is usually showing you a sodium adduct.

This is the version I would want a new member to read first.

0 likes in reply to #35 10mo
JD
j.delacroixTL3Regular18 Sep 2025 · edited#37

This follows post #35 rather than contradicting it.

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

A qualification I should have led with rather than closed on.

3 likes 10mo
IB
i.brobergTL222 Sep 2025#38

That matches what I have seen, for whatever a single anecdote is worth.

10 likes 10mo
N
NorringtonTL3Regular25 Sep 2025#39
d.ferreira, post #17: Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass. I have deliberately not rounded that, because the rounding is where the argument starts. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

0 likes in reply to #17 10mo

Suggested topics

TopicParticipantsRepliesViewsActivity
About the Mass spectrometry category
LC-MS and high-resolution MS, charge states, mass accuracy in ppm, and what identity confirmation does and does not establish. This post is a community wiki: any member at trust level 3 or above can edit it,…
DOOBLDBDRV+5 9 19k 6h
Distinguishing a deletion sequence by mass alone — the long version
On the subject in the title: Distinguishing a deletion sequence by mass alone — the long version Working notes rather than a conclusion. A narrow question about Distinguishing a deletion sequence,…
TFVKPFIDLF+38 44 60k 11mo
Second pass at: Counter-ion content and its effect on measured mass
Second pass at: Counter-ion content and its effect on measured mass Writing it up because I had to work it out twice and would rather nobody else did. Something about Counter-ion content does not reconcile…
SORSLAFFOV 4 16k 4mo
Distinguishing a deletion sequence by mass alone
Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Trying to establish the boundary conditions on distinguishing a deletion…
RMLSMHMINM+121 129 25k 17mo
[2026 update] Charge states for a 4 kDa peptide, worked through
On the subject in the title: Charge states for a 4 kDa peptide, worked through Working notes rather than a conclusion. Working through the identity arithmetic and I would like it checked. semaglutide has a…
DBBIATNAL+116 135 31k 3mo

Related topics — sharing the tags LC-MS, charge states, MALDI-TOF

TopicParticipantsRepliesViewsActivity
Oxidation and deamidation mass shifts, tabulated
Posting this under the heading it deserves: Oxidation and deamidation mass shifts, tabulated Everything below is what sits behind that. Oxidation and deamidation mass shifts, and specifically the version of…
AFJVIATNCS+67 76 21k 16mo
Second pass at: Where impurities in solid-phase peptide synthesis come from
Posting this under the heading it deserves: Second pass at: Where impurities in solid-phase peptide synthesis come from Everything below is what sits behind that. Working notes on impurities in solid-phase…
SABJBPJSMS+47 51 3.3k 5mo
Residual solvents and scavengers: what is looked for
On the subject in the title: Residual solvents and scavengers: what is looked for Working notes rather than a conclusion. I would like to know what people here actually do about residual solvents and…
DOEVABKPLG+19 23 39k 14mo
Particulates in a reconstituted solution: identifying the likely source
Posting this under the heading it deserves: Particulates in a reconstituted solution: identifying the likely source Everything below is what sits behind that. Something about particulates does not reconcile…
SCNECCYMRM+15 19 13k 9mo
Isotope patterns and why the monoisotopic peak is not the tallest — does this still hold?
Isotope patterns and why the monoisotopic peak is not the tallest — does this still hold? — that is the question, and I have not found it answered plainly anywhere I have looked. Asking about Isotope patterns…
PODBTD 2 4.6k 7h