Reading an MS report that only gives you a single number posts 31–60

This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1.

CN
cannula_notesTL2Member17 Mar 2025#31

Everything in post #29 holds. The case it does not cover is the one I have.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

5 likes 16mo
AW
am.wikstromTL219 Mar 2025 · edited#32

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

I have kept the units in throughout, for the obvious reason.

0 likes 16mo
I
IbrahimoviTL2Member20 Mar 2025#33
sa.vogel, post #6: Adducts — sodium and potassium especially — shift the observed mass in predictable increments. A report showing an unexplained plus twenty-two is usually showing you a sodium adduct. Go to post

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

Not the answer, but possibly the question that gets there.

0 likes in reply to #6 16mo
ZN
z.nakamuraTL221 Mar 2025#34
cannula_notes, post #31: Everything in post #29 holds. The case it does not cover is the one I have. Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Go to post

Building on post #33 rather than restating it.

The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.

20 likes in reply to #31 16mo
D
DSakamotoTL3Regular23 Mar 2025#35

Thank you — that answers what I came here to find out.

2 likes 16mo
CK
c.kuuselaTL224 Mar 2025#36

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

Worth one more sentence than it usually gets.

0 likes 16mo
TF
taper_fileTL3Regular25 Mar 2025#37
j.habermann, post #22: Coming back to post #18, because the follow-up matters more than the original answer. Electrospray on a peptide of this size gives a multiply charged series rather than a single ion. Seeing only one charge state usually means the deconvolution has already been done for you, which is worth knowing. Go to post

Where I part company with post #33, and it is a narrow parting.

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

28 likes in reply to #22 16mo
HK
h.krastevTL227 Mar 2025#38

Post #37 is the version of this I will quote in future. One addition.

Tandem mass spectrometry with fragmentation gives sequence information that intact mass cannot. It is the analysis that would actually confirm identity, and it is rarely supplied.

14 likes 16mo
PS
p.silvaTL228 Mar 2025#39

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

On reflection I would soften that slightly.

0 likes 16mo
MM
m.malinowskiTL229 Mar 2025#40

Post #37 is right about the mechanism and I think understates the practical bit.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

If that reads as pedantic, it is, and it has saved me twice.

0 likes 16mo
AM
a.molnarTL231 Mar 2025#41

That is a cleaner way of putting what I was circling around.

2 likes 16mo
BT
baseline_tableTL2Member1 Apr 2025#42
h.brandt, post #11: Useful. I have added it to my own notes with the date on it. Go to post

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

The evidence for this is thinner than the way I have phrased it suggests.

8 likes in reply to #11 16mo
AV
a.villalobosTL22 Apr 2025#43

Picking up post #42: that is the part I would want checked first.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

One of those cases where knowing the mechanism does not help the decision.

27 likes 16mo
D
DSakamotoTL33 Apr 2025#44
JL
j.lokkenTL25 Apr 2025#45
a.molnar, post #41: That is a cleaner way of putting what I was circling around. Go to post

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

4 likes in reply to #41 16mo
TF
taper_fileTL3Regular6 Apr 2025#46
LJankowiak, post #7: Building on post #6 rather than restating it. Tandem mass spectrometry with fragmentation gives sequence information that intact mass cannot. It is the analysis that would actually confirm identity, and it is rarely supplied. I am reporting what happened, not recommending it. Go to post

The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.

I would treat that as a working assumption and revisit it.

13 likes in reply to #7 16mo
DA
d.achebeTL27 Apr 2025#47

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

Small point, but it is the one that usually catches people.

0 likes 16mo
D
DKwiatkowskiTL3Regular8 Apr 2025 · edited#48

Post #45 describes the usual case. This is about the unusual one.

An acylated peptide has a mass that reflects the modification, so comparing against the mass of the unmodified backbone gives a mismatch that is not an error.

It took me longer than it should have to see that.

0 likes 16mo
FP
f.piresTL210 Apr 2025#49
e.mikkelsen, post #24: The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had. Posted with less confidence than the sentence structure implies. Go to post

Building on post #46 rather than restating it.

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

0 likes in reply to #24 16mo
GD
glossary_deskTL3Regular11 Apr 2025#50

No disagreement from me. Posting only so the question does not look ignored.

2 likes 16mo
AP
au.pereiraTL212 Apr 2025#51

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

The right answer here may simply be that it has not been measured.

24 likes 16mo
LE
logbook_erinTL313 Apr 2025#52
FD
f.danquahTL215 Apr 2025#53

On post #51 — agreed on the reasoning, with one qualification.

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

I have seen it go both ways, which is why I hedge.

1 like 15mo
MM
maintenance_modeTL3Regular16 Apr 2025#54
a.villalobos, post #43: Picking up post #42: that is the part I would want checked first. Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. One of those cases where knowing the mechanism does not… Go to post

Picking up post #51: that is the part I would want checked first.

The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.

The short answer was in the first line; everything after is the working.

0 likes in reply to #43 15mo
PK
p.krastevTL217 Apr 2025#55

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

This is where my knowledge stops and I would rather mark the edge than blur it.

17 likes 15mo
RV
r.venkatesanTL3Wiki editor18 Apr 2025#56

Confirming post #55 from a second method, which matters more than confirming it from a second person.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

Anyone with a larger sample, please post it.

7 likes 15mo
NK
ni.kravchenkoTL219 Apr 2025#57

Post #55 describes the usual case. This is about the unusual one.

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

0 likes 15mo
IS
isotonic_sheetTL3Regular21 Apr 2025#58
p.trevino, post #13: Desalting before analysis: some samples need desalting to remove salts that suppress the peptide signal. Report whether desalting was used, because it can affect the apparent ionization efficiency and the reported purity. Go to post

Understood, and I withdraw the assumption I opened with.

0 likes in reply to #13 15mo
AV
a.vermeulenTL222 Apr 2025#59
r.oyelaran, post #21: Useful. I had the fact and not the reason, which turns out to be the important half. Go to post

Where I part company with post #55, and it is a narrow parting.

The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.

Someone should write this up properly, and it should probably not be me.

0 likes in reply to #21 15mo
RJ
r.jhannsdttirTL3Regular23 Apr 2025 · edited#60

Post #59 is the version of this I will quote in future. One addition.

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

23 likes 15mo