endpoint_marginTL2Member16 Aug 2024#31

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

The answer changed when I changed how I was measuring, which was informative.

13 likes 23mo
RC
r.coelhoTL216 Aug 2024#32

Reading back through, this was answered upthread and I missed it. My fault.

27 likes 23mo
K
KStephanopoulosTL3Regular16 Aug 2024#33
c.vasquez, post #10: Small methodological point on High-resolution: repeating a measurement is cheap and resolves most of what is being argued about here at no cost to anyone. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

Worth reading the earlier posts in this thread before acting on mine.

0 likes in reply to #10 23mo
SV
s.vogelTL217 Aug 2024#34
O
OTeixeiraTL3Regular17 Aug 2024#35

The arithmetic in post #33 is right; the assumption feeding it is the part to check.

Marking my uncertainty on High-resolution explicitly. I am confident about the direction, much less confident about the size, and not confident at all that it generalises past the case in the first post.

19 likes 23mo
SO
s.okonkwoTL217 Aug 2024#36

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

A single observation, in a thread that deserves better than single observations.

0 likes 23mo
BP
bench_peakTL3Regular17 Aug 2024 · edited#37
crossref_check, post #7: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. That holds under the stated conditions and I have stated them. Go to post

Mass error in parts per million is (observed minus theoretical) divided by theoretical, times a million. On a high-resolution instrument a low single-digit figure is unremarkable and expected.

I would put the burden of proof on the interesting explanation, not the dull one.

0 likes in reply to #7 23mo
SO
s.oyelaranTL217 Aug 2024#38
m.strand_rph, post #16: Right — I had this wrong and I am glad to have read it before it mattered. Go to post

Electrospray on a peptide of this size gives a multiply charged series rather than a single ion. Seeing only one charge state usually means the deconvolution has already been done for you, which is worth knowing.

Marking that as an opinion rather than a finding.

4 likes in reply to #16 23mo
B
BirkelandTL318 Aug 2024#39
PF
p.fontaineTL218 Aug 2024#40

Post #38 describes the usual case. This is about the unusual one.

The version of High-resolution that circulates here is a simplification of a simplification. It is not wrong, but it has lost the conditions under which it holds, and those conditions are where the interesting cases live.

0 likes 23mo
RM
r.molnarTL218 Aug 2024#41

I read post #40 twice before replying, because I had assumed the opposite.

I have no financial interest in anything named in this thread and I want to say so before I comment on High-resolution, because it is the sort of subject where it matters.

6 likes 23mo
C
CSagredoTL3Regular18 Aug 2024#42

Post #40 answers the question as asked. The question underneath it is different.

The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.

If that reads as pedantic, it is, and it has saved me twice.

1 like 23mo
HR
h.ramosTL218 Aug 2024#43
customs_ledger, post #9: Post #5 and I disagree about the size of the effect, not about the direction. Where I would push back on the High-resolution consensus is the confidence, not the direction. The direction looks right. The confidence is borrowed. Go to post

Two people in this thread mean different things by High-resolution and are disagreeing about the definition while believing they are disagreeing about the facts. Worth pausing to define it.

32 likes in reply to #9 23mo
F
FFaulknerTL3Regular19 Aug 2024#44
h.ramos, post #43: Two people in this thread mean different things by High-resolution and are disagreeing about the definition while believing they are disagreeing about the facts. Worth pausing to define it. Go to post

Electrospray ionisation produces multiply charged ions. For a 4 kDa peptide you expect mostly 2+, 3+, and 4+ charge states. Reading an electrospray spectrum means recognizing the envelope, not looking for one peak.

16 likes in reply to #43 23mo
HA
h.amankwahTL219 Aug 2024 · edited#45

Coming back to post #43, because the follow-up matters more than the original answer.

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

This is the version I would want a new member to read first.

10 likes 23mo
T
ThibodeauTL3Regular19 Aug 2024#46

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

That has been true for the cases I have seen and I have not seen many.

3 likes 23mo
FC
f.chowdhuryTL219 Aug 2024#47
c.ostergaard, post #15: High-resolution has been discussed here with more heat than it deserves, mostly because two definitions have been in play the whole time. Go to post

Genuine question rather than a rhetorical one: has anyone here actually observed High-resolution, as opposed to read about it? The thread is long and I cannot tell.

0 likes in reply to #15 23mo
CI
c.inglethorpeTL3Regular19 Aug 2024#48

Adding thanks rather than a view. I do not have a view worth the space.

23 likes 23mo
LV
l.vermeulenTL220 Aug 2024#49

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

The literature is thinner on this than the confidence in the thread implies.

15 likes 23mo
N
NardoneTL2Member20 Aug 2024#50
r.coelho, post #32: Reading back through, this was answered upthread and I missed it. My fault. Go to post

Calibration matters: a high-resolution instrument out of calibration can report mass with ppm error large enough to be uninformative. Check when the instrument was last calibrated before trusting the reported accuracy.

If this contradicts something upthread, the upthread version may well be the better one.

6 likes in reply to #32 23mo
YE
y.eriksenTL220 Aug 2024#51

My understanding of High-resolution is a few years old and may have been superseded. If it has been, I would genuinely like to know rather than keep repeating it.

11 likes 23mo
P
PSundbergTL2Member20 Aug 2024#52

An observation about High-resolution that I cannot explain and am posting anyway, on the principle that unexplained observations are more useful public than private.

24 likes 23mo
FF
f.fontaineTL220 Aug 2024 · edited#53

This follows post #50 rather than contradicting it.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

Not the answer, but possibly the question that gets there.

0 likes 23mo
RM
r.marsdenTL3Regular21 Aug 2024#54
crossref_check, post #7: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. That holds under the stated conditions and I have stated them. Go to post

Worth separating two things that post #52 runs together.

The reason High-resolution is hard to answer is that the obvious measurement and the relevant quantity are not the same thing, and substituting one for the other is silent.

3 likes in reply to #7 23mo
NS
n.serranoTL221 Aug 2024#55

Understood, and I withdraw the assumption I opened with.

7 likes 23mo
L
LeitermanTL3Regular21 Aug 2024#56

Practical experience of High-resolution, offered as one case with the conditions stated, not as a general finding. Conditions first, because they are what make it interpretable.

18 likes 23mo
CB
c.balogunTL221 Aug 2024#57

Taking post #54 at face value and following it one step further.

Deamidation adds approximately one dalton and produces a species that frequently elutes very close to the parent. It is the hardest common impurity to see chromatographically and the easiest to see by mass.

Somebody will have a better source than mine, and I hope they post it.

0 likes 23mo
KF
k.farrugiaTL3Regular21 Aug 2024#58
q.zhao_qa, post #14: The useful distinction on High-resolution is between what was measured and what was inferred from it. Both end up in the same sentence and only one of them has error bars. Go to post

Mass error in parts per million is (observed minus theoretical) divided by theoretical, times a million. On a high-resolution instrument a low single-digit figure is unremarkable and expected.

1 like in reply to #14 23mo
MK
m.kjaerTL222 Aug 2024#59
Birkeland, post #39: Trifluoroacetate adducts are common in material purified with TFA and are one reason a mass spectrum from a peptide can look busier than expected. Go to post

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

On reflection I would soften that slightly.

23 likes in reply to #39 23mo

Suggested topics

TopicParticipantsRepliesViewsActivity
Charge states for a 4 kDa peptide, worked through — the long version
Charge states for a 4 kDa peptide, worked through — the long version Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like…
HEGTMI 2 293 3mo
What a 2 ppm match establishes, and what it does not
What a 2 ppm match establishes, and what it does not — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
JVBPSOOAK+56 60 5.2k 9mo
Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below. I have spent a fortnight trying to pin mass spectrometry down…
DAAPINOSS+75 82 4k 17mo
Oxidation and deamidation mass shifts, tabulated
Posting this under the heading it deserves: Oxidation and deamidation mass shifts, tabulated Everything below is what sits behind that. Oxidation and deamidation mass shifts, and specifically the version of…
AFJVIATNCS+67 76 21k 16mo
MS/MS sequence confirmation: what it costs and when it is worth it
MS/MS sequence confirmation: what it costs and when it is worth it — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I repeat it…
SVSLSTGBLM+2 6 2.7k 1d

Related topics — sharing the tags LC-MS, counter-ion content, MALDI-TOF

TopicParticipantsRepliesViewsActivity
Particulates in a reconstituted solution: identifying the likely source — what changed since
Particulates in a reconstituted solution: identifying the likely source — what changed since — setting out what I have, and where I think it stops being reliable. What changes if the standard account of…
GPMSCV 2 129 14mo
Endotoxin and why it is a separate question from purity — what changed since
On the subject in the title: Endotoxin and why it is a separate question from purity — what changed since Working notes rather than a conclusion. What changes if the standard account of Endotoxin is wrong? I…
JSPRMVMDEM+78 91 1k 10h
Charge states for a 4 kDa peptide, worked through — the long version
Charge states for a 4 kDa peptide, worked through — the long version Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like…
HEGTMI 2 293 3mo
Coming back to: Oxidation and deamidation mass shifts, tabulated
Oxidation and deamidation mass shifts, tabulated — setting out what I have, and where I think it stops being reliable. Reading back through what has been written here about Oxidation and deamidation mass…
HFSLADAKTI+34 38 4.8k 10mo
[2026 update] Buffers, pH and why a formulation is a formulation
Posting this under the heading it deserves: Buffers, pH and why a formulation is a formulation Everything below is what sits behind that. A formulation question rather than a pharmacological one, though the…
PDRRS 2 773 3mo